Specification
2-1. ECG CMRR ¡Ã60dB Electrodes RA, LA, LL, RL, Chest (5 Leads) Leads ¥°, ¥±, ¥², aVR, aVL, aVF, V1 Frequency Bandwidth 0.5~40Hz ( case of filter ON) Display 0.5, 1, 2, 4mV/cm Input impedance ¡Ã10M¥Ø Voltage 4000VAC 60Sec, 50/60Hz Patient leakage current ¡Â100uA Filter 50/60Hz Pulse Range 30~300bpm Accuracy ¡¾1bpm
2-2. SpO2 Range 35~100% Accuracy ¡¾1%(90~100%) ¡¾2%(80~89%) ¡¾3%(65~79%) blow ¡¾4%(35~64%) Pulse Range 25~255bpm Accuracy ¡¾1bpm
2-3. NIBP Range Adult 20~300mmHg Infant 10~150mmHg Accuracy ¡¾5mmHg or 2% Max. pressure Adult 300+30mmHg Infant 150+15mmHg
2-4. Respiration Current ¡Â500uA Frequency 50KHz Method Chest impedance measurement by electrodes Range 2~99bpm Accuracy ¡¾1bpm Display 0.2~4¥Ø/cm 2-5. Temperature Range 15.0~45.0¡É Resolution 0.1¡É Probe Option
2-6. Displays Display 6.4" TFT COLOR LCD Clearance horizontal RGB 640 ¡¿vertical 480 pixel Pixel pitch 0.20¡¿0.20mm Displays real time graph (ECG, SpO2 and Resp.)
2-7. Built-in printer (option) Speed ; 25, 50mm/sec Paper width ; 58mm
2-8. Power External AC AC85~265V, 50/60Hz External DC DC Adaptor (Model : MW160) DC12~28V, 3A, Medical Safety Internal DC Rechargeable 1.2V¡¿8ea, 1.2AH 1 hour
2-9. Classification Class I / CF type
2-10. General Max. power consumption ¡Â 20Watt Connection to printer RS-232C 2ports Dimension 275(W)¡¿220(H)¡¿210(D)mm Weight 2.65Kg(without printer) Operation Condition 10~40¡É / 30~75% Storage Condition -10~60¡É / 30~75%
* Important notice
Electro Surgery Interference Suppression Protection Test
Date : 2006-05-08 Hospital : Samsung Cheil General Hospital Surgery : breast cancer Duration : 1hour Equipment :KTPM-2002 patient monitor Hewlett Packard Monitor (Agilent Component Monitoring System & V24 and V26) Electric Surgical unit
Power on electric surgical unit : 15W/ 30W/ 40W
Test Result ; 1. 15W/ 30W : A little noise is occurred on ECG base-line. There is no difference to read ECG waveform and hart rate. Comparing with HP(Hewlett Packard)monitor, there is no difference between two monitors. Doctors can read the waveform for the surgery. Amount of noise on ECG base-line is almost same with HP monitor. 2. 40W : Almost same test result with the testing on power 15W and 30W.
---From this test, we can get a conclusion that KTPM-2002 can be used with electric surgical unit during an operation.
|