Multi Parameter Patient Monitor VITAPIA 7000K Technical Specifications
| Classification |
|
| Anti-electroshock type |
Class I equipment and internal powered equipment |
| EMC type |
Class A |
| Anti-electroshock degree |
ECG(RESP), SpO2, NIBP, IBP,TEMP,CO2 CF |
| Specifications |
|
| Size moniter |
310X140x263mm(type 7012) |
| Weight monitor |
3.8K.g |
| Temperature |
Working 5~40 Transport and storage -20~-60 |
| Humidity |
Working<80% Transport and storage <80% (nocoagulate) |
| Altitude |
Working -500 to 4,600m Transport and storage -500 To 13.000m |
| Power supply |
100/240VAC, 50/60Hz PMAX= 150VA FUSE T 1.6 A (Type7012 K) 9VCD,Pmax =36PA(Type 7000H) |
| Display |
|
| Devise |
12.1" in color TFT,3LCD |
| Messages |
8Waveforms Maximum 1 alarm LCD(yellow/Red) 1 Power LCD(Green) 1 Battery charge(Yellow) 3 sound mode corresponding Alarm Mode |
| Signal Interface |
|
| External Display |
Standard VGA |
| ECG Out put |
BNC |
| Amplitude |
1mV |
| Impedance |
50 |
| Single Delay |
20mS |
| Battery |
|
| Battery |
Rechargeable 3.6A/ Hr 7.4V Li Battery |
| Operating time under the normal use and full charge greater than 120 min Operating time after the first alarm of low will be about 5 min |
| Recorder (Option) |
|
| Recorder Width |
48mm |
| Paper speed |
25/50mm/s |
| Trace |
2 |
| Recall |
|
| Trend Recall |
short : 1 hrs, 1Second Resolution long : 72hrs, 1 min Resolution |
| Alarm event Recall |
71 Alarm event of all parameters and 8/16/32 second of corresponding waveform |
| NIBP Measurement recall |
400NIBP Measurement data |
| ECG |
Lead Mode : 5 Leads(R.N.F.N.C or RA,LA,LL,RL,V) Lead selection : I, II, III, avR, avL, avF, V |
| Waveform |
2ch |
| Lead Mode |
3Leads (R.N.F.N.C or RA,LA,LL) |
| Lead selection |
I, II, III |
| Wave form |
1ch |
| Gain |
X2.5mm/mV, x5.0mm/mV, x10mm/mV, x20mm/mV |
| HR and Alarm Range Adult |
15~300 bpm |
| Neo/Ped |
15~350 bpm |
| Accuracy |
1% or 1 bpm , which great |
| Resolution |
1bpm |
| Sensitivity |
>200uVP-P |
| Differential input Impedance |
5M |
| CMRR |
Monitor>105 dB Operation >105dB Diagnosis >85dB |
| Electrode offset Potential |
300mV |
| Leakage Current |
<10uA |
| Baseline Recovery |
<3s After Defi |
| ECG signal range |
8mV(Vp-P) |
| Bandwidth |
1~20Hz(+0.4dB,-3dB) 0.5Hz~40Hz(+0.4dB-3dB 0.05Hz~75Hz(+0.4dB-3dB). 76Hz~150Hz(+0.4dB, -4.5dB) |
| Calibration Signal |
1mV(Vp-p), 5% Accuracy |
| ST Segment Monitoring Measure and Alarm |
-2.0 ~+2.0 mV |
| ARR Detection type |
ASYSOLE, VFIBNTAC,COUPLET,BIGMINY, TRIGEMINY, R ON T, VT>2, PVC, TACHY,BRADY MISSED BEATS, PNP, PNC |
| Alarm |
Available |
| Review |
Available |
| RESPARATION |
|
| Method |
Impedance between R-F(RA-LL) |
| Differential input Impedance |
>2.5 M |
| Measuring Impedance Range |
0.3~5.0 |
| Base line Impedance Range |
01.~2.5K |
| Bandwidth |
0.3~2.5Hz |
| Resp. Rate |
Adult 0~120rpm |
| Measuring and Alarm Range |
Neo/Ped 0~150 rpm Resolution 1rpm Accuracy 2rpm Apean Alarm 10~40 S |
| NIBP |
|
| Method |
Oscillometric |
| Mode |
Manual, Auto, STAT |
| Measuring interval in AUTO Mode |
1,2,3,4,5,10,15,30,60,90,120,180,240,480,760 Min |
| 5Min |
5Min |
| Plus Rate Range |
40~240 bpm |
| Alarm Type |
SYS, DIA ,MEAN |
| Measuring and alarm Range |
|
| Adult Mode |
SYS : 40~270 mmHg DIA : 10~215 mmHg MEAN : 20~235 mmHg |
| Pediatric Mode |
SYS : 40~200 mmHg DIA : 10~150 mmHg MEAN : 20~165 mmHg |
| Neonatal Mode |
SYS : 40~135 mmHg DIA : 10~100 mmHg MEAN : 20~110 mmHg |
| Resolution pressure |
1mmHg |
| Accuracy pressure |
|
| Maximum Mean error |
5 mmHg |
| Maximum standard deviation |
8 mmHg |
| Overpressure protection Adult Mode |
297 3 mmHg |
| Pediatric Mode |
240 3 mmHg |
| Neonatal Mode |
147 3 mmHg |
| SpO2 |
|
| Measuring Range |
0~100% |
| Alarm Range |
0~100% |
| Resolution |
1% |
| Accuracy |
70~100% 2% 0%~69% unspecified |
| Actualization interval |
About 1sec |
| Alarm Delay |
10sec |
| Plus Rate Measuring and Alarm Range |
0~250 bpm |
| Resolution |
1 bpm |
| Accuracy |
¢®¨ú2 bpm |
| TEMPERATURE |
|
| Channel |
2 |
| Measuring and Alarm Range |
0~50 |
| Resolution |
0.1 |
| Accuracy |
¢®¨ú0.1 |
| Actualization interval |
About 1 sec |
| Average Time Constant |
<10 Sec |
| IBP (Option) |
|
| Channel |
2 |
| Label |
ART, PA0000CVP, RAP, LAP, ICP, P1, P2 |
| Measuring and Alarm Range |
ART : -50~350 mmHg PA : -6~120 mmHg |
| CVP/RAP/LAP/ICP |
-10~40 mmHg |
| P1/P2 |
-10~300 mmHg |
| Press sensor Sensitivity |
5uV/V/ mmHg |
| Impedance |
300~3000 |
| Resolution |
1 mmHg |
| Accuracy |
2% or 1 mmHg which great |
| Actualization interval |
About 1 sec |
| CO2(Option) |
|
| Method |
Infra-red Absorption Technique |
| Measuring Mode |
Sidestream or Mainstream (Option) |
| Side-stream mode sampling gas flow rate |
50ml/min (Option) |
| Measuring Range |
CO2 : 0~150 mmHg INSCO2 : 0~150 mmHg AwRR : 0~150 rpm |
| Resolution |
CO2 : 0.1 mmHg(0~69 mmHg), 0.25 mmHg(70~150 mmHg) INSCO2 : 0.1 mmHg(0~69 mmHg), 0.25 mmHg(70~150 mmHg) |
| Accuracy CO2 |
2 mmHg, 0~40 mmHg
5% of reading, 41~70 mmHg
8% of reading, 71~100 mmHg
10% of reading, 101~150 mmHg |
| AwRR |
1 rpm |
| Initialization |
|
| Mainstream |
Capnogram displayed in less than 15sec at an ambient Temp. of 25 , full specifications within 2 min |
| Sidestream |
Capnogram displayed in less than 20sec at an ambient Temp. of 25 , full specifications within 2 min |
| Mainstream Rise Time |
Less then 60ms Adult Reusable or Single-Patient-Use Airway Adapter |
| Sidestream Delay Time |
2~3sec |
| Alarm range |
CO2 : 0~150mmHg INSCO2 : 0~150mmHg AwRR : 2~150bpm |
| Suffocation Alarm Delay |
AwRR 10~60Sec. |
|