  PRODUCT NAME : VITA 7000K 
  PRODUCT CODE :  

Multi Parameter Patient Monitor VITAPIA 7000K Technical Specifications

Classification
Anti-electroshock type Class I equipment and internal powered equipment
EMC type Class A
Anti-electroshock degree ECG(RESP), SpO2, NIBP, IBP,TEMP,CO2 CF
Specifications
Size moniter 310X140x263mm(type 7012)
Weight monitor 3.8K.g
Temperature Working 5~40
Transport and storage -20~-60
Humidity Working<80%
Transport and storage <80% (nocoagulate)
Altitude Working -500 to 4,600m
Transport and storage -500 To 13.000m
Power supply 100/240VAC, 50/60Hz PMAX= 150VA FUSE T 1.6 A (Type7012 K)
9VCD,Pmax =36PA(Type 7000H)
Display
Devise 12.1" in color TFT,3LCD
Messages 8Waveforms Maximum
1 alarm LCD(yellow/Red)
1 Power LCD(Green)
1 Battery charge(Yellow)
3 sound mode corresponding Alarm Mode
Signal Interface
External Display Standard VGA
ECG Out put BNC
Amplitude 1mV
Impedance 50
Single Delay 20mS
Battery
Battery Rechargeable 3.6A/ Hr 7.4V Li Battery
Operating time under the normal use and full charge greater than 120 min Operating time after the first alarm of low will be about 5 min
Recorder (Option)
Recorder Width 48mm
Paper speed 25/50mm/s
Trace 2
Recall
Trend Recall short : 1 hrs, 1Second Resolution
long : 72hrs, 1 min Resolution
Alarm event Recall 71 Alarm event of all parameters and 8/16/32 second of corresponding waveform
NIBP Measurement recall 400NIBP Measurement data
ECG Lead Mode : 5 Leads(R.N.F.N.C or RA,LA,LL,RL,V)
Lead selection : I, II, III, avR, avL, avF, V
Waveform 2ch
Lead Mode 3Leads (R.N.F.N.C or RA,LA,LL)
Lead selection I, II, III
Wave form 1ch
Gain X2.5mm/mV, x5.0mm/mV, x10mm/mV, x20mm/mV
HR and Alarm Range Adult 15~300 bpm
Neo/Ped 15~350 bpm
Accuracy 1% or 1 bpm , which great
Resolution 1bpm
Sensitivity >200uVP-P
Differential input Impedance 5M
CMRR Monitor>105 dB
Operation >105dB
Diagnosis >85dB
Electrode offset Potential 300mV
Leakage Current <10uA
Baseline Recovery <3s After Defi
ECG signal range 8mV(Vp-P)
Bandwidth 1~20Hz(+0.4dB,-3dB)
0.5Hz~40Hz(+0.4dB-3dB
0.05Hz~75Hz(+0.4dB-3dB). 76Hz~150Hz(+0.4dB, -4.5dB)
Calibration Signal 1mV(Vp-p), 5% Accuracy
ST Segment Monitoring Measure and Alarm -2.0 ~+2.0 mV
ARR Detection type ASYSOLE, VFIBNTAC,COUPLET,BIGMINY,
TRIGEMINY, R ON T, VT>2, PVC, TACHY,BRADY
MISSED BEATS, PNP, PNC
Alarm Available
Review Available
RESPARATION
Method Impedance between R-F(RA-LL)
Differential input Impedance >2.5 M
Measuring Impedance Range 0.3~5.0
Base line Impedance Range 01.~2.5K
Bandwidth 0.3~2.5Hz
Resp. Rate Adult 0~120rpm
Measuring and Alarm Range Neo/Ped 0~150 rpm
Resolution 1rpm
Accuracy 2rpm
Apean Alarm 10~40 S
NIBP
Method Oscillometric
Mode Manual, Auto, STAT
Measuring interval in AUTO Mode 1,2,3,4,5,10,15,30,60,90,120,180,240,480,760 Min
5Min 5Min
Plus Rate Range 40~240 bpm
Alarm Type SYS, DIA ,MEAN
Measuring and alarm Range
Adult Mode SYS : 40~270 mmHg
DIA : 10~215 mmHg
MEAN : 20~235 mmHg
Pediatric Mode SYS : 40~200 mmHg
DIA : 10~150 mmHg
MEAN : 20~165 mmHg
Neonatal Mode SYS : 40~135 mmHg
DIA : 10~100 mmHg
MEAN : 20~110 mmHg
Resolution pressure 1mmHg
Accuracy pressure
Maximum Mean error 5 mmHg
Maximum standard deviation 8 mmHg
Overpressure protection Adult Mode 2973 mmHg
Pediatric Mode 2403 mmHg
Neonatal Mode 1473 mmHg
SpO2
Measuring Range 0~100%
Alarm Range 0~100%
Resolution 1%
Accuracy 70~100%2%
0%~69% unspecified
Actualization interval About 1sec
Alarm Delay 10sec
Plus Rate Measuring and Alarm Range 0~250 bpm
Resolution 1 bpm
Accuracy ¢®¨ú2 bpm
TEMPERATURE
Channel 2
Measuring and Alarm Range 0~50
Resolution 0.1
Accuracy ¢®¨ú0.1
Actualization interval About 1 sec
Average Time Constant <10 Sec
IBP (Option)
Channel 2
Label ART, PA0000CVP, RAP, LAP, ICP, P1, P2
Measuring and Alarm Range ART : -50~350 mmHg
PA : -6~120 mmHg
CVP/RAP/LAP/ICP -10~40 mmHg
P1/P2 -10~300 mmHg
Press sensor Sensitivity 5uV/V/ mmHg
Impedance 300~3000
Resolution 1 mmHg
Accuracy 2% or 1 mmHg which great
Actualization interval About 1 sec
CO2(Option)
Method Infra-red Absorption Technique
Measuring Mode Sidestream or Mainstream (Option)
Side-stream mode sampling gas flow rate 50ml/min (Option)
Measuring Range CO2 : 0~150 mmHg
INSCO2 : 0~150 mmHg
AwRR : 0~150 rpm
Resolution CO2 : 0.1 mmHg(0~69 mmHg), 0.25 mmHg(70~150 mmHg)
INSCO2 : 0.1 mmHg(0~69 mmHg), 0.25 mmHg(70~150 mmHg)
Accuracy CO2 2 mmHg, 0~40 mmHg
5% of reading, 41~70 mmHg
8% of reading, 71~100 mmHg
10% of reading, 101~150 mmHg
AwRR 1 rpm
Initialization
Mainstream Capnogram displayed in less than 15sec at an ambient Temp. of 25, full specifications within 2 min
Sidestream Capnogram displayed in less than 20sec at an ambient Temp. of 25, full specifications within 2 min
Mainstream Rise Time Less then 60ms Adult Reusable or Single-Patient-Use Airway Adapter
Sidestream Delay Time 2~3sec
Alarm range CO2 : 0~150mmHg
INSCO2 : 0~150mmHg
AwRR : 2~150bpm
Suffocation Alarm Delay AwRR 10~60Sec.


   